Since the detection is at 375 nm, you have a lot of freedom to select your buffer system. It is reasonable to assume that the rate of acid-induced degradation is directly proportional to [H+]; in other words a factor of 10 for each pH unit. You can use formate (pH 3-4) or acetate (pH 4-5) buffers for your mobile phase and diluent to reduce the decomposition of the sample.
Mark Tracy
Senior Chemist
Dionex Corp.

I got the peak shoulder problem solved by using Phenomenex brand column instead of Zorbax column.
Thank you for all your helps.

Is the Phenomenex <5µ? That was going to be my suggestion - that you could use the same dimensions for column jacket with a smaller particle size. If what you were seeing was poor resolution of cis-trans isomers, that should help.
Thanks,
DR
Image

I have seen this kind of affect before i.e. changing just the column brand having a significant impact on the baseline of your gradient.

Are there any rational reasons for this that anyone could suggest :? ?!?

DR,
I did not have cis-tran separation problems. My problem was my reference standard peak had a shoulder (please see my first post). I got it fixed by changing column brand only. It worked! :D . But I don’t understand why it worked. :?

Thanks,
20 posts Page 2 of 2

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Google [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry