- Posts: 17
- Joined: Wed Mar 28, 2007 8:47 pm
- Location: CA
http://www.sepsci.com/chromforum/viewto ... t=perylene
Running 8270 on ocean sediment samples, on two different Agilent GC/MS systems, the recovery of the 6th and final internal standard compound, perylene-d12, will fluctuate significantly with respect to the response of chrysene-d12, the 5th internal standard.
The ratio of chrysene-d12 to perylene-d12 will vary from 1.2 to over 6. This diminishing recovery of perylene-d12 results in very high calculated values for compounds associated with this internal standard compound.
Here are some important points to consider:
These are internal standards, added after extraction and concentration, used to correct results for solvent evaporation, etc. The ratio of our target compound response to the internal standard response is compared to the same ratio in the calibration. They are not surrogate standards that go through extraction to monitor extraction efficiency.
This problem occurs both before and after cleaning source, cleaning the inlet, changing liners and liner type.
The final conclusion in the thread mentioned above is that it may have something to do with the acetone used in ASE extraction, which is what we are doing. However, the problem occurs in calibration verification standards prepared using only methylene chloride, no extraction and no acetone.
The problem seems to get worse as the run persists, suggesting that as the inlet or column gets dirty the problem gets worse. But this is not always the case.
Any ideas?