Anybody know how to go about validating a kjedahl method
ie ICH guidelines , accuracy linearity etc or not applicable here
simple intermediate precision with 2 analyst perhaps ?
Advertisement
Discussions about sample preparation: extraction, cleanup, derivatization, etc.
No matter what it is used for. It may be ISO 17025 accredited and in this case has to be validated.Why are you trying to validate the Kjeldahl method? As I recall it is used for feeds, foods, and water (TKN or Total Kjeldahl Nitrogen)!
It's not an impurity method, so I don't think you'd need to have your linearity cover that far. You could probably get away with the typical 80-120%. If you want an expanded range, 50-150%. Duplicate, triplicate, or a single curve is up to you and your internal SOPs.Linearity - weigh 30-120% of typical sample as detailed in method
prepare in duplicate ?
Yes, that would be how we did it. Alternatively per ICH Q2(R1), you can determine accuracy using a known reference standard so if you can find one where the percent nitrogen has been determined, you can show that you get the same result (or reasonably close based on analytical variability) using this method.Accuracy - 80,100, 120% level ? amounts of acetanilide ?
in triplicate ?
I do believe that Leucine is popular for shaking down a Kj analysis procedure and associated equipment (steam distillation/titration combination device). It is neither the easiest nor the hardest of the naturally occurring AAs to shake a nitrogen atom out of, so it is a decent test. Also, it is available in highly pure, well characterized gram quantities. The manual that accompanies your Kj apparatus should provide a useful range over which they consider the thing to be accurate, so I recommend using that as your guide for setting your ± range for a given sample.It's not an impurity method, so I don't think you'd need to have your linearity cover that far. You could probably get away with the typical 80-120%. If you want an expanded range, 50-150%. Duplicate, triplicate, or a single curve is up to you and your internal SOPs.Linearity - weigh 30-120% of typical sample as detailed in method
prepare in duplicate ?
Yes, that would be how we did it. Alternatively per ICH Q2(R1), you can determine accuracy using a known reference standard so if you can find one where the percent nitrogen has been determined, you can show that you get the same result (or reasonably close based on analytical variability) using this method.Accuracy - 80,100, 120% level ? amounts of acetanilide ?
in triplicate ?