I am having issues with a new gradient RP-HPLC method, which I have been running now for the past three weeks.
I am getting a symmetrical ripple effect at approx. 33 to 36 minutes in all my blank and sample injections. Please take a look at an example of this interference at the website below.
http://tinypic.com/view/?pic=xgd5ps
The following information details my experimental and operating conditions:
Experimental:
The HPLC operating conditions used were as follows:
Injection volume: 20uL
Column temperature: 25 degC
UV detector wavelength: 225nm
Run time: 55 minutes
Flow rate of 1 mL/min using a gradient program
Time (mins) A % B %
Startup 80 20
1.0 80 20
45.0 20 80
46.0 20 80
47.0 80 20
55.0 80 20
Where A % = Buffer (0.1% phosphoric acid in water)
Where B % = Acetonitrile
Sample diluent: 30/70 (v/v) acetonitrile to purified water
The symmetrical ripple effect occurs at approx. 33 to 36 mins.
The pressure decreases steadily from 90 to 65 bars as the composition of the mobile phases changes from 80/20 (v/v) Buffer/ACN to 20/80 (v/v) Buffer/ACN over a period of 45 minutes.
At 33 mins, where the interference occurs, the composition of the lines A and B are 36% and 64% respectively. The interference ends at approx 36 mins, and the composition of lines A and B are 32% and 68% respectively.
The following was investigated:
• Column changed from zorbax SB C8 to an equivalent Zorbax Rx C8. (different lot number for silica packing)
• Changed the source of milliq water (different millipore systems, taken from different purified water loops)
• Prepared mobile phase A without the phosphoric acid to determine if the phosphoric acid is causing the interference. No improvement was seen.
• Performed 0uL injection of sample diluent to investigate if sample diluent is causing the interference. No improvement made.
• Changed to a second system (from 42 to 47). No improvement.
• Could the interference be air? Degassed with helium for 20 minutes but made no difference to chromatography.
• Premixed the mobile phases A and B to give compositions 80/20 (v/v) buffer to acetonitrile and 20/80 (v/v) buffer to acetonitrile respectively. Initial results saw an improvement in the chromatography but subsequent mobile phase preparations and system setups, the interference reoccurred.
• System washed with hot water to remove any traces of previous buffers.
• System needle was washed with 100% acetonitrile by performing 25 injections at 100uL. No improvement.
• Altered the gradient table, by holding the gradient at 33 mins at composition 36/64 (v/v) buffer to acetonitrile for an extra ten minutes to see the effect on the interference. The interference had the same profile but eluted over a longer period.
• Changed the needle and needle seat No difference made to chromatography.
• Bypassed the degasser and the interference disappeared but background noise increased after several injections. I attempted to repeat this scenario on a second system but failed. Bypassing the degasser on the second system, did not get rid of the interference.
If anyone can suggest what is causing this interference I would be most grateful to you.
Any information, tips or suggestions would be received with many thanks.
Regards,
Colette[/url]