For those specific errors, from Shimadzu:
https://faq.an.shimadzu.co.jp/faq/show/ ... in=english
(Basically check gas supply is at sufficient, at recommended values)
CARm AFC valve not completely closed:
"There is a leak inside the purge flow control valve of the AFC.
Answer
Check that the split line outlet and purge line outlet are not blinded.
If the error appears repeatedly even after checking the above, turn off the main power of the instrument, stop gas supply, and contact your Shimadzu sales/service representative."
https://faq.an.shimadzu.co.jp/faq/show/ ... in=english
Regarding cleaning the split vent line (not just changing the trap): the required frequency will depend on your analytes and matrix. This is common maintenance. Most GCs now do not have heated split vent lines (a Thermo Trace we had did, and that was an added feature we installed that was a nice addition). So any high boiling material swept out of the inlet during split injection may condense in the line prior to the trap. When the line gets gunked up enough, there can be an effect on the split vent flow which causes errors. This has happened to us more than once, so when we change the split vent trap, we also clean the line from the inlet to the trap. With Agilents, the line is just standard 1/8" tubing with standard Swagelok connectors; sometimes our lines are so fouled we just replace the line.