G1946D LCMS foreline pump error

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
My Agilent 2001-vintage G1946D/1100 system shuts down randomly during a sequence with a "S4 turn on requested with high foreline pressure" followed by a "difficulty with vacuum system gauges" and finally "S4 pressure below calibrated range". The MS module icon comes up "red" other modules' are all "green". The sequence stays in "run mode" (status icons blue) and hangs with the yellow "sequence stopping" messages when you manually intervene. I cannot get the system back to "ready" state through the Chemstation. After doing the reset procedure to get MS module communication back, it runs fine for awhile, then the error will return. I'm pumping at 0.2 ml/min using a narrow bore column, so it's not like I'm flooding the source. The foreline pressure reads about 2.8-2.9 torr on average, turbos are at 98-99% all the time. The Edwards pump has been rebuilt recently and this pressure is how it read when it came back, I've NEVER seen it run below 2 torr. Data looks really good when the system is "working", that is, before the error pops up. But I cannot reliably run a long sequence overnight! Any ideas on how to proceed? Is my foreline pressure on the "ragged edge"? Any help would be much appreciated.
No, your pump is fine. Your vacuum gauge, on the other hand, is flaky. I've have to replace several, and the foreline gauge gives me exactly that type of error. Unfortunately the G1946D is kind of like the 1999 Volvo I bought for my kids. One hell of a solid mechanical box with crappy, flaky sensors stuck all over it and it won't run without the stupid sensors.

The high vac gauge is even more fun!
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
Hello, sorry I need to revive this thread because I have a smiliar-ish problem.

Im getting the "S4 turn on requested with high foreline pressure" error but not the "difficulty with vacuum" etc
the foreline gauge shows a constant ~ 1.5 torr
The high vacuum gauge just sporadically jumps to live and shows then perfectly valid high vac values.
Im getting periodic APG remote shutdown errors as a result(?).

I dont really understand here if the foreline gauge or the high vac gauge is the problem? Are they connected on a bus and the foreline tells the high vac gauge to turn on? (which one is bad then) or i agilent software just assuming this problem? Replacing both is quite expenisve unfortunately.

Does anybod know whats the culprit here?

Thanks in advance :)
Max
3 posts Page 1 of 1

Who is online

In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry