Persistent Carryover Contamination in Thermo TRACE 1600 GC –
9 posts Page 1 of 1
We’re using a Thermo TRACE 1600 GC. recently have an issue with carryover contamination. After injecting six alkenone samples in a row, we ran a solvent (hexane) blank — but the same contamination peaks, including the alkenone peaks, still appeared. We think its sample carryover.
I understand it’s good practice to include solvent blanks between every 2–3 samples to minimize this issue. Here’s what we’ve done so far to troubleshoot:
1. Baked the column at 330°C for 15 minutes
2. Cleaned and replaced the syringe cleaning solvent vials and caps
3. Purged the column for 15 minutes
4. Conditioned the column for 15 minutes
5. Replaced the liner
Despite all this, the same contamination peaks persist. Our next step is to cut about 2 cm from both ends of the column.
This issue only started this week. last week’s sequence (which included solvent blanks) ran perfectly, and the solvent chromatograms were clean, without any contamination or alkenone peaks.
Has anyone experienced a similar problem? Any suggestions or insights on what else we could try to resolve this carryover issue?
Thank you for your time.
Did you try to change the syringe ; if you have an autoinjector ?
Dirty split trap may another source.
You may have a fragment of septum or a ferrule lying in the bottom of the injector body - whihc stayed there when yo uchanged the liner. I am ot familiar with your model of GC and so I am not sur eof the best way to remove it.
At omegle ], user safety is paramount:
Advanced moderation systems
Report functionality for inappropriate behavior
Anonymous chat protection
The Omegle experience includes:
Crystal-clear video streaming
Advanced user matching https://omeglechat.online/
No, we have not changed the syringe yet — we will give it a try. We’ll also check the split trap.
Thank you for your suggestions.
Thank you for the suggestion. Could you please explain how to clean the split trap ?
I have not done it before, so any guidance would be very helpful.
Split trap is a consumable and needs to be changed with a new one.
But if you want to refresh it yourself , remove the trap from the split vent line , remove the end- fittings , empty inside and fill it with fresh activated carbon.
Thank you for the clarification. I will check it out.
Cheers
Thank you very much for your reply, I’ll check it.
Cheers
9 posts Page 1 of 1