7693A ALS injecting excess volume

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
My 7693A ALS injector has recently started occasionally injecting about 50% larger volume that it should. We are injecting 5uL from the standard 10uL syringe. Several times now, we are seeing apparent injection of ~7.5uL. e.g. a 1000 ppb standard is suddenly 1500 ppb and the surrogate is correspondingly higher at 150%. We can see that the solvent peak is distintively fatter.

If this is normal wear and tear is there a kit for servicing the detent system that controls the syringe needle pull out?
Have you caught it in the act and seen it do this? It would be quite a large error to happen mechanically. Is it possible that there is something unreliable going on in the inlet and it's a different efficiency of vaporisation/split flow or something???
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 2 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry