I see this as a common HPLC issue; my co-worker did exactly the same working on a trace pesticide assay a few months ago - standard was in methanol, but sample was in water-methanol, and standard peaks were crappy. As Tom said, use a weaker solvent for the sample/standard such as an organic-water mix, or inject less microliters of the organic-only solution.Using a too-strong diluent causes a variety of peak shape problems because the diluent mixes with the mobile phase in the column and the peak comes to a screeching halt. The distorted peak shape is the chromatographic equivalent of a skid mark. As with a headspace, the resulting distortion can be almost anything: shoulders, tails, fronts, splits, etc. depending on the mixing dynamics of the system.