we have two different strenghts of tablets for our disso test. The first one has a standard conc of 0.02mg/ml and the other has 0.025mg/ml. We then made a 0.027mg/ml as our 100% concentration for the linearity curve, the range we used is between 40% to 140%. Can we just say that the linearity curve we established is just okay for the two standard concentrations i mentioned early? This is as per our validation requirement. Thanks.
-
- Posts: 3503
- Joined: Mon Aug 30, 2004 11:43 pm
I'm not a cGMP expert, but do have a lot of experience in the OTC pharmaceutical business. Typically, our protocols and validations call for linearity to be 80 to 120% of target, with at least five real data points.
You have established a linearity range that should cover 80 to 120% for each. I would suggest just documenting that.
You have established a linearity range that should cover 80 to 120% for each. I would suggest just documenting that.
-
- Posts: 95
- Joined: Fri Nov 30, 2012 5:18 pm
Consumer Products Guy wrote:
I'm nota cGMP expert, but do have a lot of experience in the OTC pharmaceutical business. Typically, our protocols and validations call for linearity to be 80 to 120% of target, with at least five real data points.
You have established a linearity range that should cover 80 to 120% for each. I would suggest just documenting that.
can you explain further your info? thanks
-
- Posts: 3503
- Joined: Mon Aug 30, 2004 11:43 pm
jayroseville58 wrote:
can you explain further your info? thanks
We had a product that contained about 64% alcohol, validated test procedure. Then we added a product of 70% alcohol, wanted to use the same procedure (not wanting to decrease the sample size by 10% of the higher level product). So we extended our linearity curve upwards to cover 120% of the higher level product when sampled at the highest allowable amount in the test procedure, then did our precsion/accuracy, etc on the new product.
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science