I am making a 10,000 ppm stock of 2-CHLOROBIPHENYL
The 2-CHLOROBIPHENYL is a crystalline powder and is >99% pure. I am treating the powder as if it is 100% 2-CHLOROBIPHENYL

If I want a 10,000 ppm solution I was told to weigh out 0.5grams of the 2-CHLOROBIPHENYL powder into a volumetric flask and fill the flask to 50 ml with methanol. a weight/volume measurement. This doesn't seem to account for the density of methanol being less than water though.

My question is would it more accurate to weigh out 0.5g of the 2-CHLOROBIPHENYL into 49.5 grams of methanol? Total weight would be 50.0 grams, and the 2-CHLOROBIPHENYL would be at a concentration of 10,000 ppm.
I think you mean 49.5 g of methanol though. Otherwise, yes, this seems more accurate given the density of methanol.
That's the problem with "ppm"; it's ambiguous. Do they mean wt/vol or wt/wt? you need to check the details of the method (or whoever gave you the procedure) to verify which one was intended.

For solutions, much better to specify wt/volume (e.g., in this case 10 g/L).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Blazer wrote:
I think you mean 49.5 g of methanol though. Otherwise, yes, this seems more accurate given the density of methanol.



Sorry that was a typo. You are correct it should have been 49.5g and not 49.95g
tom jupille wrote:
That's the problem with "ppm"; it's ambiguous. Do they mean wt/vol or wt/wt? you need to check the details of the method (or whoever gave you the procedure) to verify which one was intended.

For solutions, much better to specify wt/volume (e.g., in this case 10 g/L).


The method only specifies to spike each sample with 5ul of 10,000ppm TFB.

Since the stock of TFB used in the curve was made using wt/volume I will do the same for this stock, I just wasn't sure if this was standard practice when making stock solutions to calculate ppm using the density of water rather than whatever the actual solvent density is. I would think this would be a problem for many applications.
"ppm" is not a proper unit - in this case it could mean micrograms per gram, micrograms per ml, (implausibly) microlitres per ml, or micomoles per mole, each of which will give you a different composition.

All the confusion would be eliminated by specifying the composition of solutions in SI units.

Peter
Peter Apps
It seems to me that since you're adding 5 µL of this standard at the end, you'd really want to know how many µg if the chlorobiphenyl you're really adding to the sample. Wt./vol makes more sense to me. That's how I do it as well.

The density of methanol doesn't vary that much at room temperature. I found data online that indicate:

10 °C = 0.801 g/mL
20 °C = 0.792 g/mL
30 °C = 0.782 g/mL

I don't know about you but my lab temperature never varies this much. Mostly, I'm around 20-23 °C. Your 49.5 g of methanol translates to:

10 °C (50 °F) = 61.8 mL
20 °C (68 °F) = 62.5 mL
30 °C (86 °F) = 63.3 mL

Even for this wide swing in temperature, you're only off by about 1.2% (61.8/62.5 = 0.989, 63.3/62.5 = 1.013). I doubt seriously that you'll ever experience this type of swing in temperature unless you're operating outside.

0.5 g/50 g x 10^6 = 10,000 µg/g
0.5 g/62.5 mL x 10^6 = 8,000 µg/mL

But if I'm adding 5 µL of this solution to sample, and I want to know that I'm adding 50 µg of the chlorobiphenyl to that sample, I'd go with 0.5 g chlorobiphenyl to 50 mL of solution with methanol as the dilution. The actual mass of methanol that you're adding at 20-23 °C doesn't vary a whole lot (<< 1.2%).

Just my $0.02.
7 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry