How to change the RT display on chromatograms
I am using a Chemstation software. Is there a way to change the GC retention time decimal places (displaying 3 decimal places on the chromatogram for retention time and would want to change to one decimal place)
You want the RTDECIMAL function;

RTDECIMAL #

Available only in Data Analysis. Set the decimal place labeling of integrated chromatograms between 0 and 5.

(You'll need to refresh afterwards for the current chromatogram, but it should be good after that)
How to change the RT display on chromatograms
I am using a Chemstation software. Is there a way to change the GC retention time decimal places (displaying 3 decimal places on the chromatogram for retention time and would want to change to one decimal place).
MSD ChemStation E.02.02.1431.
Thanks,
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry