Agilent GC/MSD can't conduct a run
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
3 posts
Page 1 of 1
The Agilent GC/MSD (7890A, 5975C) pumps down and tunes just fine. When I try to start a run with a manual injection, I get either "GC run aborted" error or "1101,"({SA00} :RUN:RDY ON;): Mass spec fault during acquisition: 16384" error. I've tried reinstalling the software, updated the firmware with no success. Any suggestions?
-
- Posts: 1
- Joined: Tue Oct 31, 2017 10:29 am
Hello,
i have the same problem with my GC/MSD 6890/5975 . Have you solved this problem?
In logbook are this error messages
1101,"({RUN} :MSC:RDY ON): Mass spec couldn't be turned on due to fault: 16384",1201,"({RUN} acq:strt sim): Mass spec not ready for acquisition"
-310,"({SA00} :MSC:RDY OFF;): APG RCI controller hardware not connected or not present."
-310,"({SA00} MASK 4;): APG RCI DATA OVERWRITE error occurred."
Thank´s
i have the same problem with my GC/MSD 6890/5975 . Have you solved this problem?
In logbook are this error messages
1101,"({RUN} :MSC:RDY ON): Mass spec couldn't be turned on due to fault: 16384",1201,"({RUN} acq:strt sim): Mass spec not ready for acquisition"
-310,"({SA00} :MSC:RDY OFF;): APG RCI controller hardware not connected or not present."
-310,"({SA00} MASK 4;): APG RCI DATA OVERWRITE error occurred."
Thank´s
-
- Posts: 348
- Joined: Mon Sep 13, 2010 8:28 am
Maybe APG start/ready cable connected loosely (9-pin D-DUB connector).
As far I know there is stub cable to make MS detector believe the system is ready and allow it start. Try to start w/o gc or replace cable.
Also you can try with tweezers (remove cable from GC side) short the ready signal to shield and short the start signal to shield to begin analysis.
As far I know there is stub cable to make MS detector believe the system is ready and allow it start. Try to start w/o gc or replace cable.
Also you can try with tweezers (remove cable from GC side) short the ready signal to shield and short the start signal to shield to begin analysis.
3 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Majestic-12 [Bot], pacerlaser and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Majestic-12 [Bot], pacerlaser and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science