Last week my lab experienced a hard drive failure for the workstation that runs a Trace Ultra, Polaris Q, Triplus AS system. This system has been running without significant issues since early 2013. After installing a new workstation and installing Foundation and Xcalibur 2.2 SP1 along with the drivers for the three components, I am having issues with the handshake settings between the the Trace and the Triplus. If I configure as specified in the manuals the Triplus does not inject at beginning of the run and Trace starts the run anyway. By changing the handshake settings on the Trace keypad I can get the Triplus to inject, but the run timer actually starts a full minute before the actual injection as the solvent pre-injection wash cycles are completed. No matter how I try to play musical chairs with the handshake settings I cannot get this system to properly initiate the run at contact closure.
Xcalibur shows proper communication with all three devices.
Any ideas on how to resolve this issue would be greatly appreciated.