Dear all,

I am facing a big problem with my LC-ESI-QQQ.
I am running a multi-method with ~20 analytes. All of them exhibit linear calibration curves over the working range.

Accidentally I measured 2 samples that were very high in 4 of the analytes.
Since then, the calibration curves for those compounds are "flat" (no slope), i.e., all concentrations give more or less the same signal.

I first thougt of a contamination somewere in the LC-system, as the blanks were a little compromised, however, after cleaning the system, the blanks are clean, still the calibration standards give the same response for any concentration (10 to 70 µg/L).

Can there be a phenomenon such as a memory effect for those ions? all other analytes behave normal...

Did anybody experience this?

Thanks
Georg
Perhaps the problem is not at the LC side. Look at the signals for those particular before and after the injection of high concentration: is the baseline elevated? It is possible that those analytes are being detected continually, and just give a slight bump when they elute.

I would start with performing normal ESI maintenance: cleaning the cone and the chamber.

For troubleshooting, you could manually check the MS signals when there's a flow of mobile phase over the column. Check if you can find the compounds like when you're tuning, but without introducing them.
Thanks for the hint.

The baseline was elevated before HPLC maintainance, now the baseline is acceptable.

The signal for any standard (independent of its concentration) is about the response for the medium standard (40 µg/L) of the calibration before the problem, however, the blanks are clean... looks weird.

Anyway, I will do a extended ESI-source and ion focusing unit maintainance, maybe it helps.

BR
Georg
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry