Phosphoric acid in lcms
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
19 posts
Page 1 of 2
I am developing a method for vitamin b1 and b6 that contain phosphate groups. The peaks show bad tailing. One way of geting the tailing out is ad phosphoric acid to the elluens to compete with the phosphate in my compound in adsorption. I know phosphate-bufers are not recomended om lcms but what about pure phosphoric acid?
-
- Posts: 129
- Joined: Sat Nov 30, 2013 8:44 am
- Location: Belarus
In general non-volatile buffers are not OK. Boiling point of H3PO4 is about 200 °C but I am not sure…
Best regards,
Dmitriy A. Perlow
Dmitriy A. Perlow
-
- Posts: 43
- Joined: Wed Aug 09, 2017 5:17 pm
Bp of phosphoric acid is 158 celsius, acatic acid ks 118.
i know about the buffer,, that is why i asked about phosphoric acid. by the way, my highest calibator is 0,5 umol/l so 0,5 mmol/l phosphoric acid will give me a ratio of 1000 in phosphoric acid vs analyzed compound. i think that would be enough for the compettion in adsorbtion.
i know about the buffer,, that is why i asked about phosphoric acid. by the way, my highest calibator is 0,5 umol/l so 0,5 mmol/l phosphoric acid will give me a ratio of 1000 in phosphoric acid vs analyzed compound. i think that would be enough for the compettion in adsorbtion.
-
- Posts: 184
- Joined: Mon Feb 22, 2010 11:15 am
From the literature it is usually recommended to avoid phosphate buffers and phosphoric acid because of their low volatility. I am not a good chemist but it seems to me that boiling point and volatility (or vapor pressure) are two different things. Otherwise fog would not bother me in the winter. There are some good chemists on this forum who will certainly be able to expand on this.
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
- Location: Moscow, Russia
You can use low levels of phosphoric acid but the ion source must be cleaned more frequently (daily basis etc.). Have you tried TFA etc.?Aldijkhuizen wrote:
I know phosphate-bufers are not recomended om lcms but what about pure phosphoric acid?
-
- Posts: 43
- Joined: Wed Aug 09, 2017 5:17 pm
Vlad,
i am a litlebit scared to use TFA. the lcms system is used for 6 different aplications and even more in the near future. i hear scary stories from people not geting TFA out of the system once the used it.
i am a litlebit scared to use TFA. the lcms system is used for 6 different aplications and even more in the near future. i hear scary stories from people not geting TFA out of the system once the used it.
-
- Posts: 404
- Joined: Wed Nov 09, 2016 12:43 am
Instead of phosphoric acid for protonation you can use 0.1% formic acid on your LC/MS since it is volatile.
-
- Posts: 43
- Joined: Wed Aug 09, 2017 5:17 pm
I tried formic acid but it gave very bad peaksbape die ons compound
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
- Location: Moscow, Russia
The phosphoric acid is much more "scary" than TFA I never had troubles with TFA (0.01 - 0.10 %v; of course some trace will remain, but i never had issues with that). On the other hand what choice do you have in this situation? IP-RP-HPLC? HILIC or Mixed-mode HPLC (may be, it's best choice)?Aldijkhuizen wrote:
i am a litlebit scared to use TFA. the lcms system is used for 6 different aplications and even more in the near future. i hear scary stories from people not geting TFA out of the system once the used it.
-
- Posts: 43
- Joined: Wed Aug 09, 2017 5:17 pm
I ame doing mixed mode. That is the only way i ame geting enough retention. The peak broadening is there in every mode even when i ame bypassing the column. Accept when the ph is 9.5 or higher but that is above limit of column and ionexchange probably wont work
-
- Posts: 43
- Joined: Wed Aug 09, 2017 5:17 pm
So tfa is mostly scrarry if you do full scan ms and not really a problem if you do mrm?
-
- Posts: 404
- Joined: Wed Nov 09, 2016 12:43 am
In that case you will probably get bad peak shapes with TFA too. You will have to determine the pKa of these molecules so only 1 species is present during HPLC (protonated versus ionized).
This is the reason behind protonation via formic acid, phosphoric acid, TFA.
This is the reason behind protonation via formic acid, phosphoric acid, TFA.
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
- Location: Moscow, Russia
Suppression of ionization is main issue with TFA.Aldijkhuizen wrote:
So tfa is mostly scrarry if you do full scan ms and not really a problem if you do mrm?
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
- Location: Moscow, Russia
May be pH > 9 is only way to go (a lot of sorbents works fine in this range).Aldijkhuizen wrote:
The peak broadening is there in every mode even when i ame bypassing the column. Accept when the ph is 9.5 or higher...
Also try to search some info:
https://www.agilent.com/cs/library/post ... 202015.pdf
http://www.sielc.com/wp-content/uploads ... 2017-1.pdf
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC4489891/
-
- Posts: 43
- Joined: Wed Aug 09, 2017 5:17 pm
Sav,
The firts link is about thiamine, not thiamine pyrophosphate. I had seen the third link. I think the first peak comes to early so i like to avoid this if posible. The second link is the aplication vlad from sielc developed for me. I ame trying to reproduce this but get horible results with b1 on my agilent 1290.
The firts link is about thiamine, not thiamine pyrophosphate. I had seen the third link. I think the first peak comes to early so i like to avoid this if posible. The second link is the aplication vlad from sielc developed for me. I ame trying to reproduce this but get horible results with b1 on my agilent 1290.
19 posts
Page 1 of 2
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science