Phosphoric acid in lcms
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
19 posts
Page 2 of 2
There have been several posts on the use of TFA in LC-MS on this forum. If your MS must be used for several versatile applications and by different users, avoid TFA unless you don't mind having a few enemies in your lab. TFA persistence in the MS is very long even if you clean it thouroughly and ion suppression in negative ionization is almost total.
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
- Location: Moscow, Russia
TPP is there (see table 2 and chroms).Aldijkhuizen wrote:
The firts link is about thiamine, not thiamine pyrophosphate.
It's not an issue if you will can validate method. I've successfully developed 3 methods even without column (FIA-MS; Flow Injection Analysis).Aldijkhuizen wrote:
I think the first peak comes to early so i like to avoid this if posible.
I think that your LC system contain some adsorption sites which cause TPP peak broadening (hypothesis).Aldijkhuizen wrote:
The second link is the aplication vlad from sielc developed for me. I ame trying to reproduce this but get horible results with b1 on my agilent 1290.
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
- Location: Moscow, Russia
Some people just use TFA as FA (etc.) and make mistakes (configuration of LC system, wash solution, wash sequence, methods dev sequence etc.). We used several methods at single LC-MS (+-ESI) without issues within 3 years (two methods contained 0.05 - 0.025 %v TFA).Gaetan Glauser wrote:
There have been several posts on the use of TFA in LC-MS on this forum.
I agree that it's much more safely to avoid TFA when system already contain several validated methods and operators.
-
- Posts: 1859
- Joined: Fri Aug 08, 2008 11:54 am
For avoiding TFA, in some applications where 0.1% formic acid isn't quite enough to improve peak-shape, you can use more formic acid without disaster. I have a regular method that uses 1%, and I've tried it to 2% before I got too scared and went back down again. I have a colleague who went up to 10% but I vaguely remember he dissolved part of his HPLC...
19 posts
Page 2 of 2
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science