Dear all!

We have to measure BTEX, THM, trichloroethylene etc. and 2-butoxyethanol from air. We have a Markes TD-100 7890-7000C GC-MS system. We tried out Tenax and Carbograph/Carboxen/Tenax trap with Tenax and graphitised carbon cold traps. The TD method was: 6 min desorb time (300 °C), the cold trap: -20°C then 320 °C with 3 min hold time. The used column was PONA. The TD is directly connected to the column.

We have only problem 2-butoxyethanol it gives a broading peak about 0,5-0,7 min wide peak. The other compounds gives really nice peaks, we tried out even free phenols and with a little tailing but they give really nice peaks. We tried to leave out TD from the system and directly inject to the column with an SL inhlet and it was really good without TD.

Thank u for any helps!
Erdno
erdno wrote:
We have only problem 2-butoxyethanol it gives a broading peak about 0,5-0,7 min wide peak.
Catalytic thermodegradation? Are you use inert tubes (specially deactivated) for TD?
Dear Sav!

We are using inert tubes and caps and the flow path should be inert too (as i know). I tried to decrease the temperature (120°C for flow path and 200 °C for desorption temp) but it didnt use too.

Erdno
I presume you're using a 5-phase column. Please correct me if that is not the case. Butyl cellosolve tails pretty badly on my 5-phases. Perhaps you could go to a wax or a cyano phase and get better chromatography.
rb6banjo wrote:
I presume you're using a 5-phase column. Please correct me if that is not the case. Butyl cellosolve tails pretty badly on my 5-phases. Perhaps you could go to a wax or a cyano phase and get better chromatography.

Seems it is not a problem of inappropriate column since he wrote:
"We tried to leave out TD from the system and directly inject to the column with an SL inhlet and it was really good without TD."
Split or splitless?
Our TD is directly connected to do column with an union, so no split on inlet. At TD after the cold trap we use 20:1 split (tried out 80:1 split too).

Erdno
What if you use Tenax only in the cold trap?

Or the carbon only?

The 2-butoxyethanol may be adsorbing too strongly in the cold trapping material. At -20C can you trap these analytes on something inert like just glass beads or the walls of the trap itself?
The past is there to guide us into the future, not to dwell in.
Dear James!

I tried out Tenax cold trap and carbon cold trap too. Both of them was same. I dont have any empty trap to try out if i can trap them on wall (or glass beads).

Erdno
erdno wrote:
the flow path should be inert too (as i know).
Try to replace transfer line (if you can). I don't think that its desorption issue since phenols are ok.
10 posts Page 1 of 1

Who is online

In total there are 2 users online :: 0 registered, 1 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry