Shimadzu QP2010s Problems
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
6 posts
Page 1 of 1
My lab has a Shimadzu QP2010s MS trouble with declining surrogate recovery that the Manufacturer repair rep said was too much water. The water peak has about as much intensity as the nitrogen peak. Is this normal? ...and if not, how can I get the water down?
-
- Posts: 12
- Joined: Fri Jan 29, 2016 1:07 am
In the tuning menu, you can use "peak monitor view" to check for leaks. Target m/z = 18, and then you can start troubleshooting. Ideally there should be no peak at all when you do this. The ferrules on both ends of the column are common spots, as is the o-ring on the inlet.
-
- Posts: 2
- Joined: Fri Jun 03, 2016 5:20 pm
The Shimadzu has a 2010 GC and an O-I Analytical Eclipse 4600 P&T. We've changed the column, had O-I replace the P&T. The rough pump and turbo were replaced and only 6 months old. I don't think the system has any leaks because the vacuum indicator is normal. I know O-I has a reputation for poor water management but I think it might be something in the MS. Thanks very much for any help.
-
- Posts: 12
- Joined: Fri Jan 29, 2016 1:07 am
If you plug the detector end with a blank ferrule (no column), do you still see a water peak?
-
- Posts: 5
- Joined: Sun Aug 07, 2016 2:52 pm
Hi.
If you´re watching the peak of water the same height as the nitrogen you should plug the MS entrance with a blank ferrule, put a hight temperature in the interface 250ºC perhaps and then let it pump down overnight.
That sure can help. If not, check the O-ring in the front door and replace it with a new one if neccesary.
If you´re watching the peak of water the same height as the nitrogen you should plug the MS entrance with a blank ferrule, put a hight temperature in the interface 250ºC perhaps and then let it pump down overnight.
That sure can help. If not, check the O-ring in the front door and replace it with a new one if neccesary.
-
- Posts: 1
- Joined: Sun Jan 07, 2018 11:00 am
Hi.
I have a problem with this model.
how can I connect to one of you that have this model.
tnx
I have a problem with this model.
how can I connect to one of you that have this model.
tnx
6 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science