The Fault 8 No Emission Current is usually a blown filament but can be any other reason that it won't power the filaments. I can't remember if on the 5970 there is a switch for EI/CI or if that was introduced on the 5971, but if it is set to EI and it senses too much load on the filaments when it tries to fire them up, it will assume that the pressure in the analyzer chamber is too high and shut down the filament to prevent it from burning out. If it has he switch you can flip it to CI and it bypasses the check, but I am pretty sure that is only on the 5971(it has been 20 years since I had a 5970).
Other question is why would you turn off the turbo pump when not using the instrument? They don't pull nearly as much current as the rough pump so it is best to just leave them on all the time and warmed up so the vacuum stays at optimal levels. Spinning the turbo pump up and down wears them out faster than just leaving them running full speed.