- Posts: 8
- Joined: Mon Feb 29, 2016 5:21 pm
i have a "problem" regarding an LC-MS/MS method for pegylated acrylate monomers. I started method dev with water/methanol/formic acid on a ESI-QQQ. All analytes that are not charged by itself are ionized as [M+Na]+ only, no [M+H]+ or other adducts. adding ammonium formate did not improve the situation, as i got a 7/3- mixture of sodium and ammonium adduct and traces of [M+H]+. I want to get rid of the Na-adducts as i observed that there is only very little fragmentation and SRM-transitions are weak, thus LoD is bad.
I started to add Li-formate (10 µM) which works like a charm, however, a non-volatile salt in ESI-MS
As I extended the calibration range of the method (was 0.5-100 ng/mL) to 500 ng/mL I realized that calibration curves start to flatten out at the top end and I think this is due to Li-ion shortage (10 µM = 500 ng/mL for a MW 500 compound) and there are 50+ analytes present (MW 150 to 1500).
What do you think is a concentration of non-volatile salt which is still "safe" to handle? Can I go to 100 µM?
Does anyone have experience with LC-MS and non-volatile buffers/salts?
Thanks
Georg