by
lmh » Mon Mar 27, 2017 2:32 pm
I'm puzzled. HPLC chemist, I'm not sure if you're referring to an internal standard used in calibration of a quantitative method? I'm guessing that the original poster is referring to a sodium formate calibration solution being used to provide mass calibration of an accurate mass instrument?
It is certainly possible to publish accurate mass MS data where calibration is by inclusion of a short infusion of sodium formate at some convenient point in the run, and this infusion is used to calibrate the remainder of the run. I think Bruker used to do it that way (I have no idea what they currently recommend). The question is how fast your MS drifts. If it drifts very fast, then calibrating using an infusion from some minutes previously wouldn't be as reliable as using continuous alternation of lock-mass and sample, in the way that Waters would. The important thing for publication is whether your mass accuracies remain within acceptable bounds - usually 5ppm.
The remaining question is whether it's OK to infuse sodium formate manually before quickly reattaching the tubing, rather than doing this automatically. Since you're not altering the pre-column plumbing in any way, the retention times cannot be changed by your activities. In an academic environment I think it would be grossly unfair of a referee to refuse your paper for doing this manual infusion, but you should be honest about it.
It might be easier to automate this system than you think. Do you have a divert valve on the MS? If so, you can plumb it to do loop injection, with a fairly big loop. For most of the run, you keep the loop out of the flow-path, and during this time you can fill the loop with sodium formate, either manually or by a very large syringe pump working at a very low flow. At a convenient blank bit of baseline, either before your retained peaks, or during re-equilibration of the column, switch to the "inject" position, so that the flow takes the loop's contents to the MS. The switching is, of course, automatic.