Hello James,
Thank you for the suggestion - I will definitely try it.
I would like to bring your attention to one more issue with the instrument. With each sequence we have retention time shifts to the left.
I am looking at PCB analysis in SPME extracts and have a nice MRM method with quite large number of groups for this purpose.
Whenever running a sequence with extracts and standards and QA/QCs, with time I have retention time shifts such that the peaks a partially cut off from the group.
This is causing troubles with compound quantification. After all, I do need to adjust the time of specific groups before next analysis.
I am not sure if this is caused by the GC - potentially a leak which I am not able to track or due to dirty samples. The instrument is checked for leaks with a Restek leak detector and any external leaks at the inlet or to the MS transfer line are always eliminated. This is also confirmed with the air leak check.
Even though we analyze SPME extracts, the extraction lens and repeller are covered with a purple and brown layer and I do need to clean the source more often.
Still, not convinced that it would cause retention time shifts.
Please note, we have a new column but the same story happened with the old column as well as with the new one.
Any suggestions are appreciated
M