I'm setting up Agilent 6210 TOF LCMSD. after fixing vacuum problems, I'm going to next step.
when I turn on the MASS, vacuum starts from 1×10-5 and after 30 min reaches 1×10-6 and after 7 hours reaches 5×10-7. I think Vacuum is OK for now.
Now,I've some errors and questions and will be really thankful if somebody helps me solving them.
1-what should I do after vacuum reached 5*10-7? should I start HV condition? what should I do, or I should not do?
2-I've error on status telling me:"Ion Optic slicer fault". this error comes up when vacuum reaches 2*10-6. I don't know what can cause this and what should I check. Please let me know how can I deal with this error.
3-Another warning is about Ion source.it seems that my ESI source could not be identified and error comes: Ion source is not identical".
4- I've run diagnostic tool, and got some error messages:
masked fault: Apci heater sensor open
masked fault : RF fault
masked fault: mass dac calibration fault
masked fault: iris fault
I'll be really happy if you help me solving this problems and getting this instrument up and working.
PS: if you've any detailed service manual for this instrument, I'll be thankful to have it.