Sievers 255 NCD detector
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
Everything seems to be ok, the vacuum pump and everything else, but there are random peaks that show up even when a sample is not injected. Could this be contamination? Or perhaps an electrical issue?
-
- Posts: 404
- Joined: Wed Nov 09, 2016 12:43 am
I would suspect electrical. Put a surge protector between the electrical outlet and the NCD.
https://www.agilent.com/cs/library/user ... -90006.pdf
https://www.agilent.com/cs/library/user ... -90006.pdf
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
- Location: PA
Always look at the shape of the 'random peaks'. If they are Gaussian, it is unlikely to be electrical. If you have Gaussian peaks, they have come from the column or inlet.
Gasman
Gasman
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science