I misunderstood what the OP was desiring to do until I saw this post.Consumer Products Guy wrote:
...Once we got 4-place electronic analytical balance (I pushed for this, we had the first in the whole company) we were able to weigh much quicker, so just used a dropper into a volumetric flask. Sometimes, we would actually add some solvent to the volumetric flask, tare that, then add some sample...
I do this to dilute my Internal Standard Surrogate mix. The mix comes in methanol in a 1mL vial but I never get 1mL out. I have a spreadsheet set up. I weigh and then tare a clean 40mL VOA vial. Then I add about half of the mass I expect I will need of P&T methanol as a "keeper' and record that weight. Then I tare again and crack the vial and add the contents and get that weight into the spreadsheet. I tare again and add methanol untill I reach the weight my spreadsheet tells me, having kept track and calculated the amount needed to get the target ppm. I like to dilute my 2000 ppm ISS stock to 70 ppm with this method. 5uL in my 5 mL gastight syringe is 70 ppb at the instrument. Since I split at 40:1, that would be 8.75 ng for my BFB.