We are using an Inficon Micro GC fusion. We are trying to confirm that trace nitrogen in ultra-pure Hydrogen is lower than 2ppm. We have no issues with peaks when we use 100ppm-1000ppm nitrogen in balance Hydrogen, however when we are measuring down in the 0-10ppm range we have peaks in the negatives. We have tried everything in terms of troubleshooting: We have done a bakeout, verified that there are no leaks, our carrier gas is 99.9995% pure, our fittings are tight, contamination is not an issue, our inlet psi is right at 60, and ensured that the machine is sufficiently warmed up.
The negative peaks we are seeing are consistant but we are not sure if it is safe to use the data we have to verify our purity.
we run the tests at:
inject ms:35
column temp profile: 50 C (10 sec)- 100 C (2 degrees/30 sec)
heat injector: 50 C
column pressure: 25psi
sample pump time: 30 seconds
Thank you,
Frustrated Technician