Agilent 7890B GC Species Detection Limit
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
I have an Agilent 7890B GC with model #: G3440B. It has a FID and two TCDs. I'm wondering how I can find out what the detection limits are for the gas species that I can measure. For example, I'm pretty sure the detection limit for hydrocarbons using the FID is 5 ppm. However, I'm not sure what the detection limits are for oxygen, nitrogen, carbon dioxide, carbon monoxide, helium, etc., are using the TCDs. Any help would be appreciated.
-
- Posts: 5328
- Joined: Thu Oct 13, 2005 2:29 pm
- Location: Maun Botswana
You run a series of standards of known concentration, the detection limit is somewhere between the concentration for the smallest peak you can see, and the next one down. More formally you need to find what concentration gives a signal: baseline noise ratio of 3:1.
Peter
Peter
Peter Apps
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science