Split flow set point issues

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
Comrades,

I am experiencing issues on an Agilent 7890 GC-FID and hoped for some insight before digging into the split vent line and trap.

Conditions:
split/splitless inlet running in split mode
column flow 4 ml/min
split ratio 5
split flow 20 ml/min
total flow 27 ml/min


When the method is initialized or "set up" in Empower and the instrument tries to go to the set points, in order to reach the 20 ml/min split flow set point, the column flow jumps to around 8 ml/min. If the column flow is manually set to the correct 4 ml/min column flow on the GC keypad, it drops the split flow to 6 ml/min.

No leaks could be identified using electronic leak detector. I have narrowed down the issue in my mind to restriction in the split vent line or restriction in the split trap, but I wanted to get some input before I changed these out because they are a pain in the *ss to get to. I am hoping something more sinister isn't at work here like a malfunctioning EPC.

Any recommendations appreciated as always.

Regards,

Mike
Hi, Did u mention the dimension of the column in the GC method preparation.....???
Oops - 15 m column .53 id ; column length verified on GC keypad
Have you checked the flow at the split vent to verify that it is at 20ml/min when you set the flows manually from the GC keypad?
Are you selecting constant pressure or constant flow ?

When the column flow is 4 ml/min, what is the inlet pressure ?

Have you measured the column flow with a flow meter or dead time, or are you trusting the GC readout ?

Check that you have selected the right carrier gas.

Perter
Peter Apps
I would check my column configuration on the GC panel and make sure col 1 and col 2 are not configured for the same inj or det. I have had similar problems when both col 1 and col 2 were configured for the same inj or det. I have only had this problem with the 7890 and Empower 2 and 3 not for the 6890.

ps I thought I posted this reply yesterday but don't see it. If it's a duplicate I apologize.
6 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry