Oxidation of ECD
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
Hi, can you help me, please. I have problem with GC. Is it possible that oxidation of ECD happen due to lack of N make up during the anylisis.
-
- Posts: 348
- Joined: Mon Sep 13, 2010 8:28 am
ECD machines staying spare can catch atmospheric O2 and yes, the detector loose sensitivity.
As far I remember Gurus recommend heating it high (up to 350) and conditioning up to 8 hours. If possible - switch from N2 to H2 (to remove oxidation), obviously be careful because in that case H2 is not burned.
As far I remember Gurus recommend heating it high (up to 350) and conditioning up to 8 hours. If possible - switch from N2 to H2 (to remove oxidation), obviously be careful because in that case H2 is not burned.
-
- Posts: 2
- Joined: Wed Sep 20, 2017 3:13 pm
Dear Anton, many thanks for your reply.
You've helped.me resolve my dilemma.
Many thanks once again
You've helped.me resolve my dilemma.
Many thanks once again
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science