Splitting of Hydrogen peak in GC TCD analysis!!!!!!!

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi,

I observed a splitting of hydrogen peak during a GC TCD analysis.The other peaks observed were for air ans carbon dioxide and were sharp continuous peaks. I am using a capillary column for GC and He is the carrier gas. Can someone reason out the cause for this split? Thanks

Jayasree
Which column are you using? I presume the hydrogen is at a high concentration, so you can see it?
Where can I buy the kit they use in CSI?
You are probably in the proper concentration for the infamous 'W' peak shape for hydrogen when using helium as a carrier. If your detector polarity is reversed it becomes a 'M' peak.

Change your carrier to nitrogen and see if the peak suddenly becomes one peak, and manifold times larger!

best wishes,

Rod
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry