lemonleslie
Forgive my soapbox response, but after dealing with issues like yours with management, the solution became clear and undeniable. Never use external standards for quantifying analytes by headspace.
It takes a little longer (not always, do you ever run multiple preps of ext stds?) but it removes all issues of linearity and accuracy, and sample homogeneity if you use standard addition methods to measure your impurities.
Simply, make three preparations of your sample.
Add your standard solution once, twice, and thrice (or not at all). Add the standard solvent to the same three preparations, twice, once, and none, (or three times).
Example: 100 mg of drug weighed out into each of 3 vials.
Add 100µL of Standard solution: solvents (1mg/mL of each) in DMF
to first vial.
Add 200µL of Standard solution to second vial.
Add 300 µL of Standard solution to third vial. (or add 0µL of Std Sol.)
Then add 200µL of DMF to first vial.
100 µL to second vial
and 0µL of DMF to third vial (or add 300 µL of DMF to third vial)
Now run the 3 preparations. Say this takes 1 hour (20 minutes each)
if you ran one sample and one Std by external measurement it would take 40 minutes, and another 20 minutes arguing whether the results were good or not.
But using standard addition method, you immediately will demonstrate linearity of the method and homogeneity of the sample. Two important concerns about which you will never have solid answers if running only one preparation and one standard.
Plus if the results are non-linear you know you must determine the cause of the problem, but at least you KNOW you have a problem. Running an external std method will NOT give you that information.
Auditors love to see SOLID and UNDENIABLE science when they review research. Standard addition methods avoid so much controversy.
I advocate that you always use this technique unless you are sample limited. (been there, done that when I had to)
best wishes,
Rod