Anyway, I'm trying to run residual solvents <467> by headspace with our 6890 and 1888 and I'm having a couple problems.
First, some background: We are looking for specific solvents (methanol, acetonitrile, toluene, DMF) in pyrantel pamoate; dilute with DMSO. I'm trying to stay as true to the USP as possible, to avoid running a method validation (not sure that will hold true much longer). I recently had the whole system contaminated by shoddy workmanship on our new gas manifold (see the thread titled "Ghost Peaks"). Set-up: ZB-624, 30m x 0.53mm x 3um. 35cm/min He. Oven: 40C for 10min, 10C/min to 240C for 20min. Inlet @ 160C, FID @ 250C. Headspace: 45min equil. @ 85C, 1 min pressurization @ 10psi, 1mL loop @ 95C, transfer line @ 125C.
Now, the problems:
1. All my peaks are tiny. I know, they're residual solvents, they're not supposed to be huge. But only 3-4pA, 5-10 area counts? That seems a little too small.
2. I'm getting a lot of carryover on the DMF peak. I talked to Agilent about this, they suggested increasing inlet and transfer line temps (from 140 & 105 to 160 & 125). This is running over the weekend, but first glance tests aren't looking too good.
3. My peaks seem to be tailing. Not sure if this is just because of size? DMF tails the worst.
4. My work has a web block, so I can't post pics of chromatograms. Sorry.
Thanks for any help in advance.