I work for a flavor company that produces savory flavors. We make both encapsulated and simple plated flavor powders.
Loss on drying works very well for encapsulated flavors. I just take a moisture balance to 85 degrees C which is high enough to drive off the water but low enough not to cause sugars to react or decompose.
The second category the simple plated flavors are more troublesome. Upon heating they lose flavor volatiles along with water giving inaccurate inflated readings for moisture. I am looking into Karl Fischer but they contain sugar polymers and a lot of salt which will not solubilize and many contain aldehydes, diketones, mercaptans, siloxanes all which are known KF interferrants.
I am wondering if a HS-GC-TCD system might be the simplest solution. The setup isn't much more than a KF and there are likely no solvents. The trouble I am wondering is if I heat up the sample enough to release all the water and avoid headspace matrix effects will it decompose and produce water and how to deal with the issues.