- Posts: 12
- Joined: Fri Dec 02, 2011 9:55 pm
Agilent headspace vial shaking
Discussions about GC and other "gas phase" separation techniques.
7 posts
Page 1 of 1
Quick question: we are looking at headspace units and our methodology would involve a significant shaking/mixing time (at least 30 minutes, since we are reacting in the vial and measuring one of the gaseous products.) We are looking at the Agilent 7694E and 7697A (111 vial unit with shaker.) Could anybody who uses either of these instruments tell me whether they can shake all the vials in the oven at the same time, so then we can better stagger our runs? I asked the sales rep and he did not give me a clear answer.
-
- Posts: 1108
- Joined: Wed Dec 22, 2010 10:17 pm
- Location: Chicago
I know the Tekmar 7000 only shakes one at a time though it heats up to 12 in the main carousell. I suspect the 7694 which is a similar instrument also does that.
For shaking and heating of multiple vials simultaneously you probably need either the Combi-Pal, EST Flex, or Gerstell unit.
For shaking and heating of multiple vials simultaneously you probably need either the Combi-Pal, EST Flex, or Gerstell unit.
-
- Posts: 12
- Joined: Fri Dec 02, 2011 9:55 pm
We've looked at both of the Tekmar units (Versa and HT3). Both only shake one vial at a time. The Perkin Elmer Turbomatrix will shake a dozen vials simultaneously, it's just not competitive on price.
-
- Posts: 1108
- Joined: Wed Dec 22, 2010 10:17 pm
- Location: Chicago
Actually you are in luck with the 7694 the entire carousel which holds 6 vials shakes and is within the oven.
http://ms.fiu.edu/instruments/service-i ... manual.pdf
http://ms.fiu.edu/instruments/service-i ... manual.pdf
-
- Posts: 74
- Joined: Fri Jul 18, 2014 8:22 pm
MSCHemist wrote:
Actually you are in luck with the 7694 the entire carousel which holds 6 vials shakes and is within the oven.
7694E differs from other 7694, as it is Economy type sampler, with only one space in oven. The same goes with 7697, as there are two types: one with 12 Vials/1 oven space, and the other with 111 vials / 12 oven space.
-
- Posts: 74
- Joined: Fri Jul 18, 2014 8:22 pm
I also fond this for 7697A. Agilent part number for 111/12 is G4557A for 12/1 is G4556A
http://www.chem.agilent.com/en-US/produ ... tails.aspx
http://www.chem.agilent.com/en-US/produ ... tails.aspx
- CE Instruments
Get a Dani Master SHS 120 vial positions with an 18 position agitation oven, compatible with all makes of GC. Dani made the Agilent headspace systems for years so it is their best product.
Sold by Gentech in the States and us in the UK.
Sold by Gentech in the States and us in the UK.
7 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science