I´sorry for spamming for help in my first thread but i have a problem on my headspace analysis.
I use an agilent G1888 Headspace with Agilent 6890 GC
I use a stock solution of:
Methanole: 213µg/ml
Ethanole:1000µg/ml
Isopropanole: 1000µg/ml
I dilute 400µl Stock solution with 4600µl Water for chromatography to get my calibration standard
The calibration standard is a mixture of
18µg/ml Methanole
80µg/ml Ethanole
80µg/ml Isopropanole
The Headspace Programm is:
Oven Temp: 90°C
Loop Temp: 120°C
Transferline Temp: 100°C
Cycle Time 23min
Vial Equilib Time: 10min
Temp Equilib Time: 1min
Pressurisation Time:0,2min
PressEq Time: 0,3min
Loop fill time Time: 0,15min
Loop Equilib Time: 0,5min
Inject Time: 1min
Valve Purge Time: 10min
Vial Size: 20ml
Shaking: High
GC Parameters:
Inlet:
Split Ratio: 7
Inlet Temp: 180°C
Column:
I use a Wcot fused silica column from Varian:
lenght: 30m ; Filmthickness: 1,8 µm ; Diameter: 320µm
FID Detector:
Temp: 260°C
Makeup flow (N2): 45ml/min
H2 Flow: 40ml/min
Air Flow: 450ml/min
Oven:
40°C for 8 minutes then 60°C/min to 200°C hold for 2 minutes
I get areas in my standard solution chromatogramm:
Methanole:0,9pa/min
ethanole: 7pa/min
isopropanole: 12 pa/min
after that i inject a vial filled with 5ml Water for Chromatography:
There i find:
methanole0,009 pa/min
ethanole: 0,06pa/min
isopropanole: 0,05pa/min
this is 1% carryover...
i read in your forum that your Systems have a carryover of 0,003% how can i get to that?
should i reduce the split or something?