Since DMDS and DMS as well as Methyl Mercaptain are the likely impurities, a column with basic characteristics should be picked to remove or reduce these sulfur based methanol peak interferences.
A Carbowax suitable for amine analysis would be the best choice.
If you were trying to remove methylamine (as I confused DMSO with DMF previously) then an acidic phase such as SPB-1000 would be a good choice.
Here is the reference. Search for "headspace" and "Rodney George" using Google Scholar and you will find it.
Anal. Chem., 69 (11), 2221 -2223, 1997. 10.1021/ac970040u S0003-2700(97)00040-1
Copyright © 1997 American Chemical Society
Analysis of USP Organic Volatile Impurities and Thirteen Other Common Residual Solvents by Static Headspace Analysis
Rodney B. George* and Preston D. Wright
Analytical Research, R&D Services, Mallinckrodt Inc., P.O. Box 5439, Saint Louis, Missouri 63147-0339
Received for review January 14, 1997. Accepted March 26, 1997.
Abstract:
Static headspace gas chromatography was investigated for the analysis of residual solvents in size-limited samples. The advantages of improved limits of detection at low ppm levels and decreased sample size requirements were realized. This methodology allows the measurement of 18 common residual solvents in 6 min using 1 mg or less of sample. Greater recovery of analytes from pure dissolution solvents without the use of salts became possible when smaller preparations of samples were combined with the use of smaller vials. Rapid equilibration of the static headspace sample preparations was an additional benefit of this modification. Optimized chromatography was developed to take advantage of the shorter equilibration time and to increase instrument productivity.
best wishes,
Rod