tekmar standard cross contamination
Discussions about sample preparation: extraction, cleanup, derivatization, etc.
4 posts
Page 1 of 1
Has anyone else had problems with cross-contamination between on-board standards? I have 2 different Tekmar autosamplers: an Atomx and an Aquatek 100. I am having the same problem on both instruments. I'm now using only 1 channel for the IS/Surr Mix and doing everything else by hand because I can't figure out how to stop it.
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
danelledees wrote:
Has anyone else had problems with cross-contamination between on-board standards? I have 2 different Tekmar autosamplers: an Atomx and an Aquatek 100. I am having the same problem on both instruments. I'm now using only 1 channel for the IS/Surr Mix and doing everything else by hand because I can't figure out how to stop it.
Do they use the fixed 1ul rotor valve like the Archon did? If they are using the Valco valve as the 1ul loop for internal standard introduction, those will wear and allow standard to leak from the vial into the loop between samples and cause cross contamination. The only remedy is to replace the vespel rotor, and sometimes even the entire valve is something had happened to make the valve score the rotor.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 2
- Joined: Wed Jun 29, 2011 3:50 pm
I don't think it is the same valves. Certainly not the same as the old Solateks. Everything is PEEK, no metal at all. There is a manifold that holds the standard dispensing valves and then a solenoid valve to introduce it to the sample loop. The contamination is before the solenoid valve. The standards are actually mixing to a degree. I just started running these instruments a few months ago, but looking back, the problem has been there as long as they have been installed. One is only about 9 months old. I'm thinking it must be a settings problem or just a design problem. I get absolutely consistent standard additions so I think the dispensing valves are doing their job.
-
- Posts: 29
- Joined: Wed Jun 06, 2012 6:44 pm
- Location: Mason Ohio
Hi danelledees,
I am an Applications Chemist for Teledyne Temar. I have not seen this on our Atomx or Aquatek 100s that we use in the Applications Lab. Our Technical Support staff has more experience with these types of issues. Please call out Technical Support staff at 800-874-2004 and they will be able to answer this question.
Best Regards,
Roger Bardsley
I am an Applications Chemist for Teledyne Temar. I have not seen this on our Atomx or Aquatek 100s that we use in the Applications Lab. Our Technical Support staff has more experience with these types of issues. Please call out Technical Support staff at 800-874-2004 and they will be able to answer this question.
Best Regards,
Roger Bardsley
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science