Agilent 1100 does not build up pressure

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
Dears

My Agilent 1100 quat pump does not build up pressure any more. With opened purge valve it produces some drops, then the last hanging drop (but not more) is drawn back from the outlet, then it pumps again some drops. Drops come out when the piston on the right side moves forward.
My first idea was the check valve. I replaced the cartridge => same symptom.
Then I exchanged all seals => same symptom.
Where can I look next?

Thanks in advance
Jörg
bunnahabhain wrote:
... My first idea was the check valve. I replaced the cartridge => same symptom.
...
Where can I look next?

I don't know for sure how 1100 pump is built, but probably there is another valve - active inlet valve. Both inlet and outlet valves have to be functional.
And isn't the line feeding the pump with mobile phase clogged ?
Take the column offline. Then pump at 5 mL/minute hot water, then Methanol, then IPA, and back to water. This will solubilize any blockages.

Finally pump water at 1 ml/minute. Do you still have a leak? If so find it (it is usually around where the column is attached).
bunnahabhain wrote:
Dears

My Agilent 1100 quat pump does not build up pressure any more. With opened purge valve it produces some drops, then the last hanging drop (but not more) is drawn back from the outlet, then it pumps again some drops. Drops come out when the piston on the right side moves forward.
My first idea was the check valve. I replaced the cartridge => same symptom.
Then I exchanged all seals => same symptom.
Where can I look next?

Thanks in advance
Jörg



You say no pressure? Even with purge closed? Does not sound like blockage to me, sounds more like air in lines above that purge. Have you checked that? Especially when you say that flow goes backwards briefly. One channel at a time run 100% that channel and 2-3 mL/min flow, watch for bubbles in the line.

Did someone else use the instrument and forget mobile phase and run it dry and not fully prime the instrument before you? Air can be hiding in those lines in a number of places.
Also watch the line of the incoming solvent where it goes into the bottom of the pump. If a bubble there moves forwards and backwards then you could have a bad Active Inlet valve, the solenoid at the bottom where the line attaches. These have a replaceable cartridge, different than the one at the top.
The past is there to guide us into the future, not to dwell in.
Thank you all for assistance

The outlet valve was the problem. I only replaced the inlet valve cartridge before. Sonicated the outlet valve this morning, now it is working again.

Regards,
Jörg
Thx for feedback. :-)
7 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry