by
Kenn » Sun Jan 28, 2018 11:17 am
adam wrote:
Hello
Has anyone had any experience injecting aqueous surfactant solutions (the surfactants are not the analytes, they are in there for another reason).
Let me explain better what I'm asking. We all know that, in reversed phase, you can generally inject larger volumes when the diluent is water, as compared to ACN or Methanol, etc. So I am wondering if one is injecting aqueous surfactant solutions, would the presence of the surfactant have the same effect as if one was injecting with an organic diluent like ACN or Methanol.
Has anyone had any experience with this? And, if so, could you kindly share what you've observed?
Thank You
Adam
No the surfactants won't act like CH3CN or methanol. Yes I have analyzed them, typically in spray solutions for residue or plant studies. the solutions sprayed in the field typically have surfactants in them to make a better carrier for the analyte. These solutions must be analyzed to confirm the concentrations in the tanks. Those concentrations are typically very high so I dilute them as much as possible and shoot as little as I can depending on sensitivity and they do not pose a problem. As far as sample composition strength though no, I would consider the surfactant fraction to be part of the aqueous because unless for some reason they added some organic to it, it is aqueous. For all practical purposes it is soapy water.
However, I'm running LC/MS/MS and don't remember if they were a problem in the UV with absorbance.