- Posts: 3
- Joined: Sat Jan 06, 2018 5:29 pm
I have a consistent problem with the waters UpLC H class systems. I've developed an impurity method and I'm looking to quantitate a known impurity at approximately 0.2% relative to sample concentration (500ug/ml). When I make an injection of diluent I get a small but significant peak at the retention time of my analyte. The pda spectra matches my analyte and is larger than my LOQ injection. I've tried washing the injector and flushing the system however the problem persists. I've had this problem multiple times in the past and have never been able to completely figure out where the problem is. I've read that a common problem is that the sample injector rotor can wear and buildup analyte that slowly blends onto the system with subsequent injections. Here are my chromatoraphic conditions
Mobile phase A: 0.1% tfa in water
Mobile phase B: 100% methanol
Gradient: 100%A for 1.5 min,92 %A at 8.5min, 80%A at 15 minutes, hold at 80% to 25 min, 40%A at 35 min, return to initial conditions at 36 and hold to 40
Column: 150mm x 2.1mm, 1.7um biphenyl
Column temp 35
Uv 214 and 271nm
Injection volume 2ul
Sample concentration: 25,000ug/ml and 500ug/ml
Standard concentration: 5ug/ml and 1ug/ml
Thanks