help with HPLC dionex sumit whit detector ultimate 3100

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I have a chromatograph sumitt dionex with a ultimate fluorescent detector 3100 with a column aclaim 120 C18 of 4.6 * 250 mm the problem I have is that when I inject a target I get peaks as if the column did not stabilize, but I stop that sequence and the reinjection that same white and white leaves I guess it is the column because I just washed the solvent deposits and the filters I do not know what else to do or if I recover the column but I'm scared that I bleed the column if I do
Hi Erick12,

More details regarding what you're trying to accomplish will help us to help you.

Data that is helpful to include: the flow rate used (so retention factors may be estimated considering the retention time(s) of the peak(s) involved--or you could note retention factor values in the post), the eluent composition, the sample/standard injection volume and the temperature at which the separation is being carried out.

If you could attach a chromatogram, that would also be helpful...for instance, it is hard to know what is meant when you say:

...when I inject a target I get peaks as if the column did not stabilize, but I stop that sequence and the reinjection that same white and white leaves I guess it is the column because I just washed the solvent deposits and the filters...


I cannot remember the member that posted this in my own paraphrased form--the more that you tell us, the better we can help you.
MattM
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry