rhaefe » Thu Sep 06, 2007 7:25 pm
													
																																																									
					 
					
						shaun78, 
where does your passivation recipe come from? Is it something you tried and it worked for you, or is it based on unequivocal study results?
 
I am asking, because I can see that you may remove some metal ions with citric acid and/or EDTA, but passivate? Usually one has EDTA permanntly in the mobile phase to control metal ions?
Then there was an article in LC.GC, June 2000 (in Dolan´s Troubleshooting) which presents evidence that nitric acid in lower concentrations can dissolve stainless (some stainless?).  Anyway,  at least a 50% HNO3 is recommended there. 
This has been discussed several times here,  maybe someone has something like a summary?
Hans,
here are some literature references you might find useful:
(1)K.E. Collins, C.H. Collins and C.A. Bertran: "Stainless Steel Surfaces in LC Systems, Part I-Corrossion and Erosion"; LCGC Magazine 18, 600, 2000
"Stainless Steel Surfaces in LC Systems, Part II, Passivation and Practical Recommendations, LCGC Magazine, 18, 688, 2000
(2) R. Shoup and M. Bogdan: "Passivation of Liquid Chromatography Components", LCGC Magazine, 7, 742, 1989
(3) US Patents: 5,997,742 (1999);
5,772,889 (1998); 6,017,457 (2000); 6,030,527 (2000)
There was an old application note for the HP 1090 dealing with stainless steel and LC separation of proteins. I do have a copy somewhere.
I always use 50 to 70% nitric acid for passivation and usually leave the UV detector in line. The nitric washes out quite nicely.
reversed phase ion-pairing chromatography of ds DNA (with triethylammonium acetate) is VERY sensitive in regard to transition metal contaminations. We found that titanium frits and passivated ss frits are best. Some PEEK frits were completely useless for ds DNA separations.
If you are interested in more info, contact me directly at