I am running a Agilent 1100 for medical cannabis. We have had very consistent retention times up until these last couple of weeks. I had a leak which caused pressure to drop and retention times to shift. Once I fixed that leak everything was back to normal. A couple days later, seemingly for no reason, the retention times shifted again. This time I was not able to determine a leak of any kind.

After changing the column filter, column and a piece of metal tubing that goes from the heater to the column, I am still seeing this retention time shift. I have taken a couple metal tubing pieces apart and re-connected them. Still nothing.

THCA was coming out just under 6 minutes and is now coming out at around 6.50. The pressure seems to be fine as it is resting at the same bar that it was at when functioning correctly. I am baffled because usually there is a correlation between the pressure and the retention time shift. This time everything looks to be running the same as before. What else could it be?
If it is a quad pump , you may check the gradient valve , if it is isocratic , then consider temperature effects.
Batch to batch variations in mobile phase can result in RT changes too.
Thanks,
DR
Image
Does your mobile phase require pH? I have seen dramatic changes in RT after even one or two units of pH past its optimum value, check that your pH meter is calibrated right. Failing that, a mixing issue seems likely.
run the old column again, if your RT is what you expect than you found the issue.
a new Column will not have have identical RTs as an old column
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry