Hello all,

I'm trying to use USP method for API verification. It's a gradient one and takes 72.1 minutes. Analytical column is 4.6-mm 25-cm; 3-mm packing L1, MP-A is 10 mM ammonium acetate and 5 mM sodium 1-heptane sulfonate, and MP-B is Acetonitrile and methanol mixture. The column and the system were equilibrated for 3 hours. At the beginning, we've got perfect peaks -Tailing factor: NMT 1.5 for both peaks-. However, after 10-15 injections, we lost system suitability; unacceptable peak shape. The column is new and used only for this analyze. Since it's USP method, we have to use it. Is there any suggestion? Regards.
More information would help:
- did the peak shape deteriorate gradually or abruptly?
- "peak shape" is a bit vague; is it simply tailing or something else (e.g., a shoulder0
- are all the peaks in the separation showing the same problem (or are some peaks still well-formed)?
- was the problem accompanied by changes in retention time?

The reason for asking all of this: if it's a "physical" problem (column headspace or partially-blocked frit) is should affect all the peaks in the separation in a similar manner, and the retention times should be unchanged. If it's a headspace, it will usually come on gradually; a partially-plugged frit will usually come on abruptly.

If it's a "chemical" problem (e.g., deterioration of the bonded phase), it is likely to affect some peaks more than others, and the deterioration is usually gradual.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Dear Tom,

Thank you for reply.

- The peak shape deteriorated gradually
- mostly shoulder
- There're five peaks in the chromatogram ( main peak -R+S-, İmp A,B,C,D and E. We observed all the same problem for all peaks
- There were no change in retention times

As you explained bellow, it should be physical problem. Using pre-column filter could resolve problem. We'll try and let you know result. Thanks :)

Also, we observed that there was precipitate in MP-A (ammonium acetate probably). Is there any tips to prepare/store MP contain ammonium acetate? We prepare MP daily and filter from 0.45 micron filter.

tom jupille wrote:
More information would help:
- did the peak shape deteriorate gradually or abruptly?
- "peak shape" is a bit vague; is it simply tailing or something else (e.g., a shoulder0
- are all the peaks in the separation showing the same problem (or are some peaks still well-formed)?
- was the problem accompanied by changes in retention time?

The reason for asking all of this: if it's a "physical" problem (column headspace or partially-blocked frit) is should affect all the peaks in the separation in a similar manner, and the retention times should be unchanged. If it's a headspace, it will usually come on gradually; a partially-plugged frit will usually come on abruptly.

If it's a "chemical" problem (e.g., deterioration of the bonded phase), it is likely to affect some peaks more than others, and the deterioration is usually gradual.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry