When using a phosphate buffer system how frequently should one flush the entire system with water (not only the piston back-chamber) in order to prevent crystalization and wearing of the spare parts?
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
I either kept the buffer flowing at a slow rate (even in a recycle mode with eluent going back into the mobile phase container) or I flushed the entire path (mobile phase reservoir forward through detector to waste line) with flush solvent made up similarly but without the buffer salts (phosphate) - every time.
-
- Posts: 4
- Joined: Mon Jun 19, 2017 1:39 pm
Thanks for the reply! So not even the column should be kept in the phosphate buffer...And at the end of the day either recirculate the phosphate or wash the entire system with water. The important issue behind this question are some ghost peaks obtained when injecting buffer through the column. They are somewhat reproductible but abnormal in my opinion.
-
- Posts: 1
- Joined: Fri Jul 07, 2017 2:08 pm
Consumer Products Guy wrote:
I either get fat burners for women here at a slow rate (even in a recycle mode with eluent going back into the mobile phase container) or I flushed the entire path (mobile phase reservoir forward through detector to waste line) with flush solvent made up similarly but without the buffer salts (phosphate) - every time.
Every time? Maybe I should be doing that too.
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
- Location: Alamo, CA / USA
This article from John Dolan's "LC Troubleshooting" series in LC-GC is almost 20 years old, but still very relevant:
http://www.lcresources.com/tsbible/26042008.PDF
http://www.lcresources.com/tsbible/26042008.PDF
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
Crevetus wrote:
Thanks for the reply! So not even the column should be kept in the phosphate buffer...And at the end of the day either recirculate the phosphate or wash the entire system with water.
Not water specifically: similar aqueous-organic mix but NO salts !
6 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science