Hello,

Could someone please help me in developing a method on triple quadrupole mass spectrometer. I have a method developed on HPLC-UV and the solvent system used was hexane/Isopropanol +0.1% triethylamine (90:10), and the analytes are basic in nature (Pka around 10). I am planning to develop a method on mass spec. To my knowledge it's difficult to see an ESI spectrum using hexane as a solvent. I did a trial on ESI and APCI by direct injection using the same (as given above) combination of solvents, but I wasn't able to see any peaks. The coulmn is compatible with solvents such as alkanes, alcohols (methanol, ethanol and Isopropanol), and acetonitrile. Could someone please suggest me to develop a method on mass spec. I would be happy to provide more information if needed. Thanks in advance for your reply.

Thanks,
Sravan Mansani.
I don't think ESI will work but the APCI should work fine. Try introducing your compound in your mobile phase without the column
at a few hundred uL/min. You should see your analyte peak.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry