Hello,

I am analyzing 05 standards solutions of creatinine , after 1/20 dilution , i inject it into HPLC (C18 column/DAD detection.Here are my results
Concentration mg/l /// Area
2,5 /// 4062,27
5 /// 8096,72
10 ///13255,34
15 /// 20784,85
20 /// 26949,78

I want to calculate the Limit of Detection (LoD), and Limit of Quantitation , can you tell me how, what data i need to calculate them?
You have the data you need for an estimate (you would have to verify by doing replicate injections at the LOD and LOQ).

Use Excel, do a linear regression and get the standard error of the y intercept (SE).
For LOD, it's 3.3 x SE / slope
For LOQ it's 10 x SE / slope
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thank you;

One last question , my standards solutions had these concentration 2,5//5//10//15 and 20 mg/l, a dilution to 1/20 was peformed before injection and the areas of each peak are :
Concentration /// Area
2,5 after dilution 1/20 => 0.125 mg/l /// 4062,27
5 after dilution 1/20 => 0.25 mg/l /// 8096,72
10 after dilution 1/20 => 0.5 mg/l ///13255,34
15 after dilution 1/20 => 0.75 mg/l /// 20784,85
20 after dilution 1/20 => 1 mg/l /// 26949,78

The limit of detection should be calculated with the concetration after dilution or with the stock standards ?
I found these results (with the concetration after dilution) :
SE: 584,7168989
Slope: 25941
LOD : 0,07438286
LOQ : 0,225402605
The limit of detection should be calculated with the concetration after dilution or with the stock standards ?
Depends on whether you want the LOD / LOQ in the original solution or in the diluted solution. Either one is OK, so long as you specify which one you're dealing with.

That said, I personally prefer to base it on the solution as injected (i.e., after dilution).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
... the only thing I'd add is that if it turns out your LOD and LOQ are way, way below your lowest standard (I haven't tried it with your figures) then you should really repeat the work with more realistic values. Otherwise they are an extrapolation and may not be terribly accurate. It's always good practice, anyway, with a LOD, to check that you really can detect a peak at the LOD.
5 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry