Advertisement

Peak on isocratic mode when injection the eluent

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hello everyone

I use an isocratic method (MeOH/water 65/36, pre-mixed and degassed in ultrasonic bath) over a Phenomenex Luna C18 column. As blank I always inject 20 µl of the eluent. What I did not expect but find reproducibly is a peak at about 2.6 min. First I thout the autosampler was defective because this peak coelutes with air injected from an empty vial (this was a coincidence I found this out).
But this peak occurs on different machines (Agilent 1100, two different Dionex systems), and with different batches of the eluent. Therefore I do not believe in the air hypothesis any more. I cannot believe that three different autosamplers are defective. Also I only see this when running the MeOH/water method.
Any idea what this peak could be?

Thank you,
Jörg

PS.: This peak appears regardless of MeOH/ratio (I tried 45/55 to 75/25).
Is this peak at t0 aka void? Then it could be due to a small pressure shock that occurs during injection or because of a slight refractive index mismatch between the eluent and the injected solvent (what's the washing liquid for the autosampler btw?).
No. void is 1.3-1.4 with this column. There is no washing liquid.

Jörg
You are probably seeing dissolved air. No suprise; it's quite common. For example:
http://www.shimadzu.com/an/hplc/support ... 37lab.html
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thank you Tom. This explains everything. I also see the retention time shift using different MeOH7H2O ratios deescribed here.

Jörg
5 posts Page 1 of 1

Who is online

In total there are 14 users online :: 3 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry