I encountered strange problem on Agilent 1100 Quart pump
My method uses 90% water and 10% methanol with C18 column. Recently I saw some bubbles coming out from MCGV valve and the pressure fluctuated between 80 bar to 120 bar (up and down, down and up) gradually. Sometimes the air bubble is moving in and out from active inlet valve (AIV).
I have replaced the MCGV, AIV and bubbles were reduced, but the bubble is still coming out after I pumped for 10 minutes. I have also changed the whole pump head (piston, seal, outlet ball valve, purge valve, frit), pump head driver motor and damper unit. I have also bypassed the degasser.
If I use single channel, say B at 100% (methanol 100%), the pressure is stable at 80 bar. The same stability goes with A at 100% (about 110 bar, constant). Once i use 90% water and 10% methanol, the ripple reading will start to go up from 1% to 12%! Both mobile phases have been replaced with new ones from brand new source in case of contamination. A new column is used too.
There are no change in method settings include compressibility, minimum strokes etc.
Does anyone have such experience before? Thanks for sharing.